Question: A gas chromatography was used for separating a mixture that contained 3 hydrocarbons (ortho-xylene, meta-xylene, and para-xylene). Internal diameter of the column was 3.9 mm
A gas chromatography was used for separating a mixture that contained 3 hydrocarbons (ortho-xylene, meta-xylene, and para-xylene). Internal diameter of the column was 3.9 mm and the length was 3.6 m. Conditions of the first trial are:
Room pressure = 695.1 mm Hg
Room temperature = 23.0 oC
Inlet pressure = 65.0 psi
Column temperature = 350 oC
He gas flow rate = 30.3 mL/min (measured using a soap bubble flow meter at the column outlet)
Conversions: partial pressure of water at 22.1 oC = 19.8 mm Hg; 760 mm Hg = 14.7 psi; absolute pressure = 273.15 oC.
| Compound | Retention Time tR (min) | Peak Width W (min) |
| ortho-xylene | 8.90 | 0.563 |
| meta-xylene | 7.83 | 0.498 |
| para-xylene | 7.36 | 0.467 |
The time, tM, for the unretained peak to elute was 1.65 min.
a) Selectivity (
) for meta-xylene and para-xylene
b) During the experiment, baseline resolution was not achieved between meta-xylene and para-xylene. Calculate the number of theoretical plates are needed to achieve resolution?
c) Calculate the corrected retention volume for ortho-xylene.
20