Calculate the amplitude of the signal pulse expected from a NaI(Tl)-PM tube combination under the following circumstances:
Fantastic news! We've Found the answer you've been seeking!
Question:
Calculate the amplitude of the signal pulse expected from a NaI(Tl)-PM tube combination under the following circumstances:
⦁ Radiation energy loss: 1.33 MeV
⦁ Light collection efficiency: 67%
⦁ Photocathode quantum efficiency: 23%
⦁ PM tube electron gain: 183,000
⦁ Anode capacitance: 88 pF
⦁ Anode load resistance: 2.2 x 105 ohms
⦁ Any other physical parameters you may need can be found in the text.