Question: A half bridge configuration is implemented with same strain gauges of gauge factor G F = 2 . 0 . The minimum strain needs to

A half bridge configuration is implemented with same strain gauges of gauge factor GF=2.0. The minimum strain needs to be measured is 10 microstrain, the maximum strain is around 3000 microstrain. (10 points total)
a. Show the steps to find the range of voltage output if the source supply is 5 V and an amplifier (shown on the right) is applied with
Rf=9.99k,R1=10*(4 points)
VGEt=GF2102VG=EeGF2=5v21010-62
V0=5010-6V1000=5010-6V
V0=05v
3000 me
VG=5v2300010-62=1510-3v
V0=1510-3v1000=15v
range is
0.05 v
15 V
b. Show the steps to find the minimum number of bits and full scale voltage required for the AD converter of the DAQ system. (Consider choses from 4,8,12,16,24 bits, and 5,10,15,24 Volts)(6 points)
Full scale voltage EFS=15-0.05=14.9v
A half bridge configuration is implemented with