Question: Design an IV level measurement system using a capacitive sensor consisting of two vertical concentric tubes, 1 0 mm and 1 2 mm in diameter.

Design an IV level measurement system using a capacitive sensor consisting of two vertical concentric
tubes, 10mm and 12mm in diameter. The maximum level height in the bottle is 200mm. Consider
0=8.81012 and r =80 for the IV solution.
C =20rL
ln ( R1
R2
)(1)
[5%] Calculate the capacitance for the full bottle.
[5%] Calculate the capacitance for the empty bottle.
[5%] Air bubbles were trapped in the sensors cavity. Determine whether the resulting IV solution
level will be reported as higher or as lower than the actual level