Question: Q3. [6 marks] the transducer specified in table below is chosen to measure a nominal pressure of 300 cm H2O. a) Find the sensitivity of
Q3. [6 marks] the transducer specified in table below is chosen to measure a nominal pressure of 300 cm H2O. a) Find the sensitivity of the transducer[1 marks] b) Estimate the magnitude of each elemental error affecting the measured pressure and its range? [4 marks] c) Calculate the overall instrument error affecting the measured pressure[1 marks]![Q3. [6 marks] the transducer specified in table below is chosen to](https://dsd5zvtm8ll6.cloudfront.net/si.experts.images/questions/2024/09/66f9055017df7_85566f9054f945c3.jpg)
Q3. [6 marks) the transducer specified in table below is chosen to measure a nominal pressure of 300 cm H20. a) Find the sensitivity of the transducer[1 marks] b) Estimate the magnitude of each elemental error affecting the measured pressure and its range? [4 marks] c) Calculate the overall instrument error affecting the measured pressure[1 marks] Table: Manufacturer's specification: Input range 20-5000 cm HO Output range 0-15 volts Excitation 15 V DC Performance Linearity error 0.65% reading Hysteresis error less 0.55% Span than Sensitivity error 0.25% of reading Thermal sensitivity 0.15%/C of span Thermal zero drift 0.05%/C of FS Temperature range 5-50 C error