Question: cm Q3. 16 marks the transducer specified in table below is chosen to measure a nominal pressure of 300 cm H.O. a) Find the sensitivity

 cm Q3. 16 marks the transducer specified in table below is

cm Q3. 16 marks the transducer specified in table below is chosen to measure a nominal pressure of 300 cm H.O. a) Find the sensitivity of the transducer|1 marks b) Estimate the magnitude of each elemental error affecting the measured pressure and its range? 14 marks] c) Calculate the overall instrument error affecting the measured pressure1 marks] Table: Manufacturer's specification: Input range 20-5000 H.O Output range 0-15volts Excitation 15 V DC Performance Linearity error 0.65% reading Hysteresis error less 0.55% Span than Sensitivity error 0.25% of reading Thermal sensitivity 0.15%C of span Thermal zero drift 0.05%/C of FS Temperature range 5-50C Crror